Help, I replaced the base unit and now the temperature is stuck at 20.5, pH is reading normal. Any suggestion.
Help, I replaced the base unit and now the temperature is stuck at 20.5, pH is reading normal. Any suggestion.
will try that when I get home, thanks
20.5 means the temp probe isn't plugged in all the way. Reseat the probe. Calibration has nothing to do with this.
You might be an engineer if...You have no life and can prove it mathematically.
This will not help, and will likely make matters worse. You cannot calibrate out a gross error such as this. Calibration is a fine-tuning process.
As zombie said, a reading of 20.5 may indicate a disconnected or improperly seated probe. It could also be a case of the pins on the probe are corroded with salt, or the wire has been damaged or a few other things.
Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.
thanks everyone, I'll check the connections
If you still have trouble after checking that the probe is fully seated, inspected the pins on the plug, etc, contact Neptune tech support.
Also, this thread has nothing to do with APEX Fusion, and does not belong in the APEX Fusion sub-forum; in the future, before starting a new thread, please choose the most appropriate sub-forum based on the topic of the intended thread from the list shown onthe forum site main page. I have moved this to a more suitable sub-forum.
Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.
will do, thanks
Thanks for all the input, I got it fixed. I had the probe wasn't pushed in all the way