    Salinity probe (swings at night)

    I've recently added kessil h380 to the sump, which turns on evening time. My chaeto growth is unbelievable, I'm having hard time remove it from the sump it's completely overgrown.
    At the same time I've noticed salinity swings at night. What would be the cause of it?

    The salinity swings starts same day I've added new light to the sump.

    Please see pictures...



    Sent from my iPhone using Tapatalk

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Does it correlate with temp?

    You might be an engineer if...You have no life and can prove it mathematically.

  3. #3
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Another thing to check is make sure the cord for the kessil is not running parallel with cond probe wire.

    You might be an engineer if...You have no life and can prove it mathematically.

  4. #4
    New User
    Join Date
    Oct 2016
    Location
    Acworth, GA
    Posts
    7
    Quote Originally Posted by zombie View Post
    Does it correlate with temp?

    You might be an engineer if...You have no life and can prove it mathematically.
    Doesn't look like it, since swings of salinity only happen at night.




    Sent from my iPhone using Tapatalk

  5. #5
    New User
    Join Date
    Oct 2016
    Location
    Acworth, GA
    Posts
    7
    Quote Originally Posted by zombie View Post
    Another thing to check is make sure the cord for the kessil is not running parallel with cond probe wire.

    You might be an engineer if...You have no life and can prove it mathematically.
    How could I check that?


    Sent from my iPhone using Tapatalk

  6. #6
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Look at where the cords physically are. If a power cord is within 2 inches of the conductivity probes wire and parallel (perpendicular is fine) then it can cause interference that will skew measurements.

    You might be an engineer if...You have no life and can prove it mathematically.

Similar Threads

  1. Salinity probe on PM2 showing temp reading without a probe installed
    By Dirks-70gal in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 2
    Last Post: 02-21-2022, 05:00
  2. Large/Sudden Salinity Swings
    By johntekin in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 2
    Last Post: 10-03-2017, 03:50
  3. Probe Swings
    By theishkid in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 19
    Last Post: 07-11-2017, 05:54
  4. PH Probe Swings
    By fmjets11 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 4
    Last Post: 07-15-2014, 14:29
  5. PH Probe Swings in Readings
    By JustinGR in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 1
    Last Post: 02-10-2013, 05:22

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •