    Conductivity probe - faulty or what?

    Hello all,
    I think there is a problem with the probe since even do completing the whole procedure to calibrate it and to reset it (disconnecting it from the base unit and running the calibration proceduce and then resetting the unit), the reading is not possible.
    I made a first attempt with the manual calibration and in the step in which the probe must be inserted in the 53,000 solution, the value was 1000 all the time, it didnīt change a single time.
    So I made an attempt for the automatic calibration and it couldnīt be finished for "Acceptable range" couldnīt be set up.
    So then I tried once again and did the whole procedure ok (resetting the probe values, resetting the Apex unit, and then manual calibration) and everything works at perfection until the time of introducing the probe in calibration solution, because again I get a value of 1000 which wouldnīt move at all.

    And even do the end message is calibration successful, it reads a salinity of 34.9 but it wonīt move from there (it doesnīt fluctuates, which is abnormal)

    Does anyone know what the heck is happening?

    kind regards

    Marco
    Attached Images Attached Images

  2. #2
    New User
    Join Date
    Feb 2018
    Location
    San Diego, CA
    Posts
    3
    Im having a similar problem. Mine wont even check off the acceptable range.

    Any suggestions?

  3. #3
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by mitchr1 View Post
    Im having a similar problem. Mine wont even check off the acceptable range.

    Any suggestions?
    Perform a manual calibration with the probe unplugged or quickly cycle through the calibration without waiting to "reset" it. Then do the normal calibration procedure and make sure the temp of the packet matches tank temp exactly.

    You might be an engineer if...You have no life and can prove it mathematically.

  4. #4
    New User
    Join Date
    Dec 2017
    Location
    Seattle, WA
    Posts
    14
    You may also want to isolate the cable’s path to the probe. I ended up having to run mine away and separate from all other electrical cables in order to keep it running accurately.

Similar Threads

  1. Help! Suspecting Faulty Temp Probe/connection
    By Jerburton in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 7
    Last Post: 02-19-2018, 15:19
  2. Conductivity Probe - Conducity Range set to Salinity (PPT or Conductivity) ?
    By omcnair in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 2
    Last Post: 12-25-2014, 19:31
  3. Help! Tmpx4 probe faulty?
    By ricknbauker in forum APEX Fusion
    Replies: 6
    Last Post: 10-30-2014, 14:42
  4. Possible faulty Orp probe
    By Trido in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 3
    Last Post: 09-23-2014, 05:06
  5. Display module blank + faulty sg probe
    By *Darren* in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 1
    Last Post: 03-23-2013, 20:37

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •