Thread: Programming ASM to use for Hamilton DO sensor?

  1. #1
    New User
    Join Date
    Dec 2015
    Location
    NL
    Posts
    4

    Lightbulb Programming ASM to use for Hamilton DO sensor?

    For one of my customers I need a more sophisticated solution than the Neptune (Oxiguard) DO sensor. I know that the ASM module (Advanced Sensor Module) is the key to other sensors. Is it possible to use the ASM for the autoclavebale Hamilton sensor that my customer uses?

    https://www.hamiltoncompany.com/prod...erm-DO-Arc-120

    Analog Interface 1 4-20 mA for DO, programmable, galvanically not isolated; pulse width modulation 5 kHz
    Operating Voltage 7 to 30 VDC max. 1 W


    Thanks in advance!

    Sander Hazewinkel
    LGem

  2. #2
    Apex User zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    7,067
    You would need a converter from 4-20mA to either 0-1V or 0-5V (the latter will be more accurate).