Thread: ATK constant solid green light

  1. #1
    New User
    Join Date
    Feb 2013
    Location
    Naples, Florida
    Posts
    7

    ATK constant solid green light

    Just set up my atk. Based on the instructions the light should be blue if everything is OK, green if its pumping, and red or flashing if theres an error. Im noticing that my unit stays green constantly. The pump is not running, and it seems to be working fine, filling up the tank as needed. However...constant green light.

  2. #2
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,463
    Is it connected to your Apex, or are you running the ATK standalone? The LED will only be blue if the ATK FMM is standalone and the ATK is not actively doing a top-off. It will be green when Apex-connected.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

  3. #3
    New User
    Join Date
    Feb 2013
    Location
    Naples, Florida
    Posts
    7
    Ok Thanks, That makes sense then. It is indeed connected to my apex. So solid green light is what I'm supposed to be seeing?

  4. #4
    Frequent Contributor iamchadster's Avatar
    Join Date
    Aug 2013
    Location
    Wilsonville, Oregon
    Posts
    2,639
    Quote Originally Posted by jcfaur View Post
    Ok Thanks, That makes sense then. It is indeed connected to my apex. So solid green light is what I'm supposed to be seeing?
    Yes that means that the FMM is communicating with the Apex.
    Chad

  5. #5
    New User
    Join Date
    Oct 2018
    Location
    Montreal
    Posts
    1
    This was helpful, thank you very much.

Similar Threads

  1. Solid green light on front of DŌS
    By brandongale in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 8
    Last Post: 03-06-2022, 18:54
  2. ATK v2 “blinking green light”
    By 3dav8r in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 2
    Last Post: 07-16-2019, 07:06
  3. ATK Green light keeps blinking
    By oseymour in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 1
    Last Post: 05-29-2018, 11:52
  4. Help! FMM ATK solid green light on, not staying in auto
    By legendracer79 in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 5
    Last Post: 09-17-2017, 15:34
  5. DOS Light stay Solid Green
    By MSBeavis11 in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 0
    Last Post: 10-28-2015, 17:24

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •