    Source for 6668 µS/cm conductivity calibratiion solution?

    I'd like to use a PM2 to measure TDS in a freshwater tank. (I know, I'll need to convert from µS/cm to get a TDS number).

    The PM2 manual says to use 6668 µS/cm calibration solution, but I can't find a source for it anywhere. I've looked at several chem supply houses, aquarium supply, Amazon... nada

    Does this stuff actually exist? ;-) Any leads would be very welcome.

    Thanks

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    You can make it with Morton's iodized salt from the grocery store and RODI water. The ratio is 0.425g salt per 100mL of RODI. If for example you used a 2L (about 2100mL) soda bottle you would need about 9 grams of salt.

    You might be an engineer if...You have no life and can prove it mathematically.

  3. #3
    Hoff
    Join Date
    Mar 2013
    Location
    Seattle, Pacific Time
    Posts
    15
    Thanks! That is one of the simplest, most helpful tips ever. :-)

  4. #4
    New User
    Join Date
    May 2018
    Location
    Melbourne, Victoria, Australia
    Posts
    6
    I'm also looking for the calibration solution 6668 µS/cm for a fresh water tank. The advice above is great but I'm in Australia and it doesn't appear we can get Morton's iodized salt here. Will any iodized salt do the job? Considering Neptune refers specifically to the calibration solution surely they make it/supply it. Any suggestion would be much appreciated. Thanks.

  5. #5
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Doc View Post
    I'm also looking for the calibration solution 6668 µS/cm for a fresh water tank. The advice above is great but I'm in Australia and it doesn't appear we can get Morton's iodized salt here. Will any iodized salt do the job? Considering Neptune refers specifically to the calibration solution surely they make it/supply it. Any suggestion would be much appreciated. Thanks.
    Whatever brand you have there that is pure salt. Avoid sea salt as that is too impure for this method.

    Sent from my SM-G965U using Tapatalk

  6. #6
    Frequent Visitor
    Join Date
    Feb 2014
    Location
    Brazil
    Posts
    88
    Hi,

    and for saltwater, how can I make calibration solution?
    I tryied 2 chemichal stores and nobody understood me...
    thanks

  7. #7
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Guilla View Post
    Hi,

    and for saltwater, how can I make calibration solution?
    I tryied 2 chemichal stores and nobody understood me...
    thanks
    Saltwater is 3.24 weight percent sodium chloride (table salt). See conductivity standard in this article.
    http://reefkeeping.com/issues/2004-06/rhf/index.php

    Sent from my SM-G965U using Tapatalk

  8. #8
    New User
    Join Date
    May 2018
    Location
    Melbourne, Victoria, Australia
    Posts
    6
    Thank you, that's great. Just one more thing Iodized vs non-iodized salt - should I use one over the other?

  9. #9
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Doc View Post
    Thank you, that's great. Just one more thing Iodized vs non-iodized salt - should I use one over the other?
    It makes no difference. You just dont want sea salt because it's less pure than typical table salts like morton's.

    Sent from my SM-G965U using Tapatalk

Similar Threads

  1. Help! No Obvious Source/Cause For Alarm
    By jdvarnon in forum APEX Fusion
    Replies: 1
    Last Post: 05-26-2020, 17:13
  2. New DOS Power source
    By gwlaws in forum Pre-Sales Questions
    Replies: 0
    Last Post: 02-15-2020, 09:18
  3. BSPP fittings source?
    By Pluto in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 14
    Last Post: 01-08-2018, 18:57
  4. Conductivity Calibration Solution
    By GonTR in forum Misc Aquarium Automation Discussions
    Replies: 4
    Last Post: 07-20-2015, 08:24
  5. Help! Conductivity Calibration solution
    By weitzel82 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 6
    Last Post: 01-30-2014, 18:04

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •