    pH probe variance? What is acceptable?

    I have two pH probes. Both have been calibrated at the same time. One probe will go in my calcium reactor, the other in my sump.

    Right now though, the probes are in the same compartment in my sump, but read different values: 8.16 and 8.28. When I noticed this, I re-calibrated both probes. No change.


    What is an acceptable variance for pH probe accuracy?

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    The maximum error should be the RMS of junction error (0.02 pH), linearity error (0.04 pH), measurement error (0.01), and buffer error (0.01). This results in an error of 0.047 for each probe.

    Under absolute worst case conditions where each probe is the polar opposite of the other, The maximum permissible difference would be about 0.1 pH. Considering the likelihood that all errors would compound oppositely in two probes, one should expect a difference of less than 0.065.

    You might be an engineer if...You have no life and can prove it mathematically.

Similar Threads

  1. Conductivity not giving green tick for acceptable range.
    By Batty in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 14
    Last Post: 03-28-2021, 19:52
  2. No acceptable range with salt probe calibration
    By jan boerlage in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 2
    Last Post: 11-29-2018, 09:24
  3. Chiller with internal t stat and 2 degree variance
    By ShadQ in forum Apex Programming for Heaters and Chillers
    Replies: 6
    Last Post: 04-20-2018, 22:34
  4. Help! ph probe acceptable range never goes green
    By Kordin in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 4
    Last Post: 04-15-2018, 16:24
  5. dual temp probes temp variance
    By wildwilly1 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 23
    Last Post: 10-09-2015, 10:23

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •