    Unhappy DOS not priming

    Hello,

    I am trying to get the DOS device to prime, but nothing is coming out. The wheels turn, but nothing. I have made sure the in tube is in fact in water.

    Also, I tried to take the the pumps off, but cannot.

    Please advise.

    Thank you,

    Don

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Make sure the compression fitting is on good. If correct, you should not be able to suck or blow on the tube when it's not running.

    Make sure the power supply is plugged in. It will turn on but won't run if that isn't plugged in.

    You might be an engineer if...You have no life and can prove it mathematically.

  3. #3
    Regular Vistor
    Join Date
    Sep 2016
    Location
    Indianapolis, IN
    Posts
    34
    Thanks that appears to be the problem.

  4. #4
    Frequent Visitor
    Join Date
    Jun 2015
    Location
    UK
    Posts
    153
    If you can, try to use the Neptune supplied tubing, heating the end in hot water before fitting and tightening the compression nut. Air ingress is a pain, but you might also find ingress from other joints (such as to your dosing containers) so it's worth paying attention here also to reduce annoying bubbles in the lines and inconsistent dosing.

Similar Threads

  1. DOS not priming
    By phat9899 in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 1
    Last Post: 01-04-2022, 06:51
  2. Help! DOS single head not dosing or priming
    By TorontoTim in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 3
    Last Post: 04-10-2021, 08:54
  3. Question: DOS priming impact on Trident Controlled Dosing
    By SuncrestReef in forum Trident Marine Aquarium Water Analyzer
    Replies: 1
    Last Post: 06-12-2019, 03:35
  4. Help! DOS not Priming I tried everything please Help..
    By tokigami in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 2
    Last Post: 11-06-2017, 15:17
  5. DŌS not priming!
    By Riley in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 1
    Last Post: 01-03-2015, 07:51

Tags for this Thread

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •