    pH notifications at high pH ranges.

    I have a question / comment for the APEX community. I have a DIY Kalk reactor with a pH probe in the reactor to let me know if the Kalk solution is saturated or if I need to add more. I want to program my APEX to send a notification if the pH in the reactor drops below 12.00 so I know to add more Kalk.

    It appears that the APEX will only allow notifications based on pH values less than 10.99 has anyone else ran into this issue. Some non DIY kalk reactors like Avast Marine have a pH probe for this purpose and I am sure people would like notifications.


    Sent from my iPhone using Tapatalk

  2. #2
    Frequent Visitor Todd's Avatar
    Join Date
    Jan 2013
    Location
    Huntsville, AL
    Posts
    583
    I recall seeing that in the past. Since you really only need a high/low indication, you might intentionally miscalibrate it, such that a true pH 12 reads 7 or 10.

  3. #3
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Above 12 causes major problems even for chemists with multithousand dollar probe setups. Above 12, alkali errors make the relationship so non-linear you can't even fake it by calibrating with 10 and 12-13 solutions. You could tell if it is noticeably below 12 by calibrating with 10 and 12 instead of 7 and 10 and set an alert at like 9 (would be roughly 11.3) but right around that 12 point readings go wacky with a pH probe.

    You might be an engineer if...You have no life and can prove it mathematically.

  4. #4
    Regular Vistor
    Join Date
    Feb 2017
    Location
    85225
    Posts
    20
    Quote Originally Posted by zombie View Post
    Above 12 causes major problems even for chemists with multithousand dollar probe setups. Above 12, alkali errors make the relationship so non-linear you can't even fake it by calibrating with 10 and 12-13 solutions. You could tell if it is noticeably below 12 by calibrating with 10 and 12 instead of 7 and 10 and set an alert at like 9 (would be roughly 11.3) but right around that 12 point readings go wacky with a pH probe.

    You might be an engineer if...You have no life and can prove it mathematically.
    I totally understand. I am not really looking for an exact reading. I have found that even without the notifications it is still very nice to know if the solution is saturated. Every kalk reactor I have used eventually gets so coated in kalk that you can’t see inside but at least the probe lets me know when it is lowish. I am just going to watch the number as it is no big deal if it is low for a couple days.


    Sent from my iPhone using Tapatalk

Similar Threads

  1. TRIDENT Test Ranges
    By jaketuff in forum Pre-Sales Questions
    Replies: 1
    Last Post: 12-05-2019, 20:21
  2. Power notifications
    By tdhight in forum Apex Programming for Dosing, ATO, and AWC
    Replies: 2
    Last Post: 12-27-2018, 03:15
  3. Question: Ranges displayed in charts
    By jhaegele in forum APEX Fusion
    Replies: 2
    Last Post: 09-12-2018, 06:31
  4. Temperature Ranges for On/Off of heaters questions
    By rbarr110 in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 1
    Last Post: 06-27-2018, 11:25
  5. Help! Too many notifications
    By jaredbcoding in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 2
    Last Post: 04-13-2017, 19:07

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •