    FMM locking up

    I have 3 FMM modules. One of them is dedicated to flow sensors with 3 x 1/2" sensors and 1 x 1" sensor. It's been running fine for 3 months, but in the past two weeks it has locked up twice. All my flow sensor readings dropped to 0 at the same time, and the FMM module showed up as Disconnected in Fusion. I disconnected the AquaBus cables and reconnected them, then the readings came back online. The 2nd time this happened, then FMM module still showed connected in Fusion, but Fusion said I have 1 new tile (it was one of the 1/2" sensors), and the LED on the FMM module was red. I again unplugged the AquaBus cables and reconnected them, then the sensors resumed their normal readings. I had to drag the 1/2" sensor tile back to my dashboard.

    My other two FMM modules (one for leak detectors & optical sensors in my sump, the other for optical sensors in my mixing station) are both working fine. Think this problematic FMM is defective?

  2. #2
    Frequent Contributor SuncrestReef's Avatar
    Join Date
    Feb 2018
    Location
    Beaverton, OR
    Posts
    3,105
    15 minutes after I reset the FMM, now it shows Disconnected again in Fusion, and a new FMM module showed up with a new AquaBus address.

    I'm opening a ticket with Neptune Support.

  3. #3
    Frequent Visitor rkpetersen's Avatar
    Join Date
    Jun 2017
    Location
    Near Seattle
    Posts
    940
    Is your module firmware current? Have you manually set each flow sensor to the proper flow sensor size, rather than using auto detect?

    Another thing you could try is using a different aquabus cable, or connecting the cable to a different other module (should make no difference, typically.)

  4. #4
    Frequent Contributor SuncrestReef's Avatar
    Join Date
    Feb 2018
    Location
    Beaverton, OR
    Posts
    3,105
    Quote Originally Posted by rkpetersen View Post
    Is your module firmware current? Have you manually set each flow sensor to the proper flow sensor size, rather than using auto detect?

    Another thing you could try is using a different aquabus cable, or connecting the cable to a different other module (should make no difference, typically.)
    Yes, all my modules and the Apex have the latest firmware. The FMM is configured manually for the 1/2" and 1" flow sensors.

    Neptune Support called me back and they think one of the flow sensors may be faulty and causing the FMM to lock up. They instructed me to unplug all the sensors and then only connect a single sensor for a few days to see if the problem recurs. Then try the next sensor and repeat. If it locks up, then they will examine the suspect sensor. This could take a couple weeks, but it's worth a shot.

Similar Threads

  1. Locking old threads
    By SuncrestReef in forum Forum & Website Info & Feedback
    Replies: 1
    Last Post: 04-01-2020, 13:42
  2. Apex Jr locking up.
    By mcnealmike in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 1
    Last Post: 11-23-2018, 09:35
  3. Can one FMM control another FMM's ports?
    By jgann in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 2
    Last Post: 10-15-2018, 11:16
  4. FMM Locking up
    By Xerses24 in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 15
    Last Post: 05-16-2018, 18:16
  5. APEX(4.31_3B14) Locking up at random times
    By jeremyee in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 0
    Last Post: 03-30-2014, 09:46

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •