I calibrated a new probe with the 53 solution last night. It was pretty close at 34.8. Then it started rising and continues to rise slowly. Any ideas what's going on?
2018-08-02_9-33-41.jpg
I calibrated a new probe with the 53 solution last night. It was pretty close at 34.8. Then it started rising and continues to rise slowly. Any ideas what's going on?
2018-08-02_9-33-41.jpg
That's normal for a new probe. It can take anywhere from a few hours to a week for the probe to "break in" before the readings stabilize (mine took 4 days). Recalibrate it once readings stabilize.