    Temp probe will not calibrate

    I have latest 2016 Apex got three ago all firmware up to date on everything and see my temp probe is not to temp. So went to setup and step by step to adjust what the temp is said ok updated. Now go back to Fusion shows old reading go to local shows old reading.
    So did calibrate temp in local updated go back to dashboards local and Fusion shows old reading.
    So what am I doing wrong?

  2. #2
    Frequent Visitor
    Join Date
    May 2014
    Location
    Portugal
    Posts
    389
    Hello,

    I had the same problem with an probe temp connected to PM1, not to APEX but it's something you can try. It's a simple test that will not take you long.

    Just redirect the wire of the probe temp. It may be suffering for parasitic noise from other power cords.

    All the best,
    Rui

  3. #3
    Regular Vistor
    Join Date
    Nov 2014
    Location
    PA
    Posts
    20
    Support I was talking to said rebooting system that worked.
    Thank you for replying

  4. #4
    New User
    Join Date
    Jun 2015
    Location
    Alhambra, CA
    Posts
    13
    I just purchased the new Apex to replace my Classic. I plugged in my Temp Probe and it showed a temp of 20.1 It never adjusted to the proper tank temp. I then calibrated it via Fusion and also through the Display Unit to the correct temp. However
    each time that it was calibrated the temp never updated and stayed at 20.1 I then attached the old Temp probe from my classic and that adjusted right away to the correct temp. I did restart and everything with the New Temp probe when i calibrated. Any idea whats going on. The temp is set to F and NOT C. Any Advice would help.

  5. #5
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by rveal23 View Post
    I just purchased the new Apex to replace my Classic. I plugged in my Temp Probe and it showed a temp of 20.1 It never adjusted to the proper tank temp. I then calibrated it via Fusion and also through the Display Unit to the correct temp. However
    each time that it was calibrated the temp never updated and stayed at 20.1 I then attached the old Temp probe from my classic and that adjusted right away to the correct temp. I did restart and everything with the New Temp probe when i calibrated. Any idea whats going on. The temp is set to F and NOT C. Any Advice would help.
    A reading near 20F is not a calibration issue. It means the probe isn't plugged in all the way or in rare cases it means either the probe itself failed or there is an internal hardware issue.

    Sent from my SM-G965U using Tapatalk

  6. #6
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,464
    A reading of 20F usually indicates that the probe is not connected. Less likely is that the temperature probe is damaged or defective.

    Ensure that it is fully seated in the proper jack - there should be a noticeable click when the locking tab engages.

    Calibration of any probe is a fine-tuning process. Calibration cannot adjust for gross errors. Never try to calibrate when the reading is so far off.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

  7. #7
    New User
    Join Date
    Jun 2015
    Location
    Alhambra, CA
    Posts
    13
    Unfortunately, I must have just got the defective one. I have tried all sorts of adjustments, of course all of this happens days before a vacation. Lucky for me I have my old temp Probe.
    Working with Neptune to figure out a replacement I hope!

  8. #8
    New User
    Join Date
    Mar 2019
    Location
    PST
    Posts
    14
    Did you ever resolve this? I just bought an Apex controller and the setup has been a NIGHTMARE...nobody said the holes were different sizes on the magnetic stand...the wifi keeps dropping...and now i plugged in my temperature probe and dashboard is still showing 20.6.....the default

  9. #9
    New User
    Join Date
    Jun 2015
    Location
    Alhambra, CA
    Posts
    13
    Quote Originally Posted by FishOrMen View Post
    Did you ever resolve this? I just bought an Apex controller and the setup has been a NIGHTMARE...nobody said the holes were different sizes on the magnetic stand...the wifi keeps dropping...and now i plugged in my temperature probe and dashboard is still showing 20.6.....the default
    Hey, no this never got resolved unfortunately. I went on vacation when this happened and when I returned I just let it go and kept my old temp gauge connected.

Similar Threads

  1. manual calibrate apex temp probe
    By james.mcclain in forum Apex Programming for Heaters and Chillers
    Replies: 1
    Last Post: 10-30-2021, 02:01
  2. Replies: 2
    Last Post: 08-15-2019, 08:05
  3. Temp probe reading 20.5 / pH probe won't calibrate
    By kio707 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 21
    Last Post: 03-07-2015, 16:37
  4. How to adjust/calibrate temp probe?
    By skyrye7 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 4
    Last Post: 07-07-2013, 07:59
  5. How do you calibrate temp probe
    By Daceman89 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 3
    Last Post: 03-07-2013, 06:47

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •