    Dos not pumping but lights are on

    Using dos connected with 1link to do awc. Has been working fine until I moved it yesterday. Everything is logging like it is working but the pumps are not pumping. The manual prime buttons are not working either. Lights are on and change to blue.
    also have another dose linked to the one not working with an aqua bus that is working.

  2. #2
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,463
    Check that the 1Link cable is fully seated at both ends.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

  3. #3
    Regular Vistor
    Join Date
    Aug 2016
    Location
    Eastern
    Posts
    15
    I’ve tried to unplug and plug both ends and still not working

Similar Threads

  1. DOS Pumping Out More Than Pumping In
    By raylinds in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 2
    Last Post: 02-06-2021, 08:24
  2. Help! DOS over pumping
    By roberttumais in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 10
    Last Post: 12-31-2018, 06:33
  3. Help! DOS not Pumping
    By wealthster in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 0
    Last Post: 04-15-2017, 06:38
  4. DOS Not pumping
    By The-Russ in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 2
    Last Post: 11-13-2016, 06:30
  5. Question: DOS Pumping Interval
    By Kendall in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 2
    Last Post: 07-01-2016, 11:48

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •