Thread: Issue with calibration of ph probe 10.0 packet

  1. #1
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12

    Issue with calibration of ph probe 10.0 packet

    Im doing the automatic calibration and when i get to the 10.0 packet the acceptable range has a red warning and wont calibrate and i end up with an error. The settling portion has a green checkmark. I have matched the temp of the packets to my tank and also tried rinsing and drying in between as well as tap it for air bubbles. Please any information would be helpful.

    Edit: When trying to do a manual calibration, both numbers settle at 700 -701

  2. #2
    Regular Vistor
    Join Date
    Nov 2016
    Location
    Orlando, FL
    Posts
    33
    Quote Originally Posted by Veetz89 View Post
    Im doing the automatic calibration and when i get to the 10.0 packet the acceptable range has a red warning and wont calibrate and i end up with an error. The settling portion has a green checkmark. I have matched the temp of the packets to my tank and also tried rinsing and drying in between as well as tap it for air bubbles. Please any information would be helpful.

    Edit: When trying to do a manual calibration, both numbers settle at 700 -701
    Interesting I am having the same problem. I am using the Neptune Calibration fluids and they just arrived from BRS. Anyone else have this issue or know the fix?

  3. #3
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12
    Im going to be passing by my LFS tomorrow and seeing if i can pick up another packet. Are you also having the same issue with the manual calibration?

  4. #4
    Regular Vistor
    Join Date
    Nov 2016
    Location
    Orlando, FL
    Posts
    33
    Quote Originally Posted by Veetz89 View Post
    Im going to be passing by my LFS tomorrow and seeing if i can pick up another packet. Are you also having the same issue with the manual calibration?
    I haven't tried manual yet. I just started over and the timer has gone down and its ramping back up. Strange... maybe the probe is bad? Is yours new?

  5. #5
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12
    Yea brand new, got it today

  6. #6
    Regular Vistor
    Join Date
    Nov 2016
    Location
    Orlando, FL
    Posts
    33
    I think its the fluid... is yours the orange Neptune packaged fluid (the 10.0)? When I tried the manual calibration it only went to 801 and settled there (should have settled in the 10xx range). SO after getting tired of waiting I uploaded the settings and left the probe in the fluid. Yup - shows PH of 8.01 when it should be 10.00 or close to it.
    Last edited by Navymedic; 10-18-2018 at 14:59. Reason: update

  7. #7
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Navymedic View Post
    Interesting I am having the same problem. I am using the Neptune Calibration fluids and they just arrived from BRS. Anyone else have this issue or know the fix?
    Unplug the probe and run a manual calibration. Then plug it back in and check that it reads near 0 for dry +-100 and reads in the ballpark of 500-700 in the calibration solution.

    Sent from my SM-G965U using Tapatalk

  8. #8
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,463
    Quote Originally Posted by zombie View Post
    Unplug the probe and run a manual calibration. Then plug it back in and check that it reads near 0 for dry +-100 and reads in the ballpark of 500-700 in the calibration solution.
    This is incorrect and completely inapplicable to pH probe calibration.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

  9. #9
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by RussM View Post
    This is incorrect and completely inapplicable to pH probe calibration.
    That is for conductivity. Didn't read closely enough to notice the poster mentioned pH.

    Sent from my SM-G965U using Tapatalk

  10. #10
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12
    Quote Originally Posted by Navymedic View Post
    I think its the fluid... is yours the orange Neptune packaged fluid (the 10.0)? When I tried the manual calibration it only went to 801 and settled there (should have settled in the 10xx range). SO after getting tired of waiting I uploaded the settings and left the probe in the fluid. Yup - shows PH of 8.01 when it should be 10.00 or close to it.

    So you still arent abke to calibrate right?

    Yea my package is the orange fluid. I also got it from brs. I'm a little ticked off because neptune support is asking me for a picture of the probe and where its plugged in... it can only go into one socket.. BRS also redirected me to neptune. Youd think they'd send me another packet of calibration fluid out of courtesy.

  11. #11
    Regular Vistor
    Join Date
    Nov 2016
    Location
    Orlando, FL
    Posts
    33
    Quote Originally Posted by Veetz89 View Post
    So you still arent abke to calibrate right?

    Yea my package is the orange fluid. I also got it from brs. I'm a little ticked off because neptune support is asking me for a picture of the probe and where its plugged in... it can only go into one socket.. BRS also redirected me to neptune. Youd think they'd send me another packet of calibration fluid out of courtesy.

    I will be calling them today. I will ask that they send a different calibration fluid, maybe this batch is bad...

  12. #12
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12
    Quote Originally Posted by Navymedic View Post
    I will be calling them today. I will ask that they send a different calibration fluid, maybe this batch is bad...

    I just came back from the lfs and bought 2 more packets. Same issue. I'm pretty sure it's the probe.

  13. #13
    New User
    Join Date
    Oct 2018
    Location
    new york
    Posts
    12
    Quote Originally Posted by Navymedic View Post
    I will be calling them today. I will ask that they send a different calibration fluid, maybe this batch is bad...

    Let me know what they say, They made me schedule a phone call for monday. Im pretty ticked off that i literally have had this for less than 24 hours and i cant even get their LAB GRADE probe to work...

  14. #14
    New User
    Join Date
    Oct 2018
    Location
    Minnesota
    Posts
    3
    I have the exact same issue. Not a good way to start with apex.

  15. #15
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Did you guys twist the BNC connector onto the port properly? That kind of behavior can occur when the probe isnt fully seated correctly.

    Sent from my SM-G965U using Tapatalk

  16. #16
    New User
    Join Date
    Oct 2018
    Location
    Minnesota
    Posts
    3
    Just double checked. Yes it is connected properly as far as I can yell

  17. #17
    Frequent Visitor
    Join Date
    Nov 2013
    Location
    Michigan
    Posts
    171
    Having same issue with new PH probe with the EL. Any update/resolution?

  18. #18
    New User
    Join Date
    Oct 2018
    Location
    Minnesota
    Posts
    3
    Neptune replaced my probe and the replacement works perfectly.

  19. #19
    Frequent Visitor
    Join Date
    Mar 2015
    Location
    San Jose,CA
    Posts
    528
    Did you remove the plastic cover from the probe?

    Are you calibrating the correct probe? pH probe is blue.

    Is the correct probe plugged into the correct port? pH port is in the middle of the stack while the ORP is at the bottom next to the conductivity and VDM ports.

  20. #20
    Frequent Visitor
    Join Date
    Nov 2013
    Location
    Michigan
    Posts
    171
    Quote Originally Posted by vhuang168 View Post
    Did you remove the plastic cover from the probe?

    Are you calibrating the correct probe? pH probe is blue.

    Is the correct probe plugged into the correct port? pH port is in the middle of the stack while the ORP is at the bottom next to the conductivity and VDM ports.
    Brand new probe out of the box with the EL bundle. EL controller only has one BNC connector for the pH probe. Yes, removed rubber tip. Temperature matched solutions with tank temp. Rinsed probe before and between solutions in RO.

    Existing pH probe calibrated just fine in same solutions.

  21. #21
    Frequent Visitor
    Join Date
    Mar 2015
    Location
    San Jose,CA
    Posts
    528
    Start a ticket with support. Make to document all the steps you have taken so far.

  22. #22
    Frequent Visitor
    Join Date
    Nov 2013
    Location
    Michigan
    Posts
    171
    Quote Originally Posted by vhuang168 View Post
    Start a ticket with support. Make to document all the steps you have taken so far.
    Support ticket opened - response from Neptune within a few hours. They asked for readings during a manual calibration cycle (and the readings while in the 10.0 solution was WAY off) - so hopefully that will lead us to a solution/resolution...

  23. #23
    New User
    Join Date
    Dec 2018
    Location
    32828
    Posts
    1
    I just had the same issue purchase my apex a week ago from Marine Depot , box came open inside and sealed outsine , i am having same issue with ph probe and just open a account here to read info .... good to know is not just me ,,,, i am brand new to Apex

  24. #24
    Frequent Visitor
    Join Date
    Nov 2013
    Location
    Michigan
    Posts
    171
    I had a support call with Neptue. In addition to rebooting the system and doing the auto calibration, they also rebooted, screen shared into my PC and connected to apex/local and were finally successful doing a manual calibration. The tech I worked with indicated that sometimes doing a calibration via fusion has just enough delay where it won't successfully calibrate. I hadn't read/tried the calibration via apex/local prior to the call with Neptune support.

  25. #25
    reefanatik
    Join Date
    Mar 2013
    Location
    houston, tx
    Posts
    48
    Quote Originally Posted by Detour View Post
    I had a support call with Neptue. In addition to rebooting the system and doing the auto calibration, they also rebooted, screen shared into my PC and connected to apex/local and were finally successful doing a manual calibration. The tech I worked with indicated that sometimes doing a calibration via fusion has just enough delay where it won't successfully calibrate. I hadn't read/tried the calibration via apex/local prior to the call with Neptune support.
    i was having the same issue and read your post. I was able to calibrate through my display module. The fusion calibration was not reading correctly, I guess it was the delay like the agent said. Thanks for posting, I was getting very frustrated not being able to calibrate a brand new probe with their solution.

Similar Threads

  1. Salinity Probe Calibration Issue
    By kayceejake in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 0
    Last Post: 06-30-2019, 11:16
  2. DOS Calibration issue
    By CRK1 in forum Apex Programming for Dosing, ATO, and AWC
    Replies: 1
    Last Post: 03-24-2019, 10:41
  3. Help! Calibration issue
    By schaaped in forum DŌS & DDR – Dosing and Fluid Metering System
    Replies: 8
    Last Post: 07-25-2015, 15:20
  4. Help! ph probe calibration issue
    By jrlamountain in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 1
    Last Post: 04-26-2015, 06:09
  5. Help! pH2 Calibration Issue
    By daves88 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 0
    Last Post: 04-03-2015, 23:51

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •