    Salinity probe stuck on "dry probe". Wont' go past and never been wet

    Hello,

    Setting up my salinity probe and it won't go past the dry the probe stage. It just times out or goes on forever. The probe has never been wet.

    Suggestions?

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Cincinnati Reefer View Post
    The probe has never been wet.
    This is the entire problem. A "dry" probe does not mean that it is completely dry. The conductivity probe should "soak" in your sump for at least 24 hours before even attempting a calibration and should then be lightly rinsed off in RODI and padded dry with a paper towel.

    Sent from my SM-G965U using Tapatalk

  3. #3
    New User
    Join Date
    Nov 2018
    Location
    Denver, CO
    Posts
    2
    I had seen you recommend this on some other posts and decided to try it. It has now been soaking for about 12 hours. I will pull it out tomorrow and dry it and try again. I will update this post with my results. Thank you for replying zombie.

  4. #4
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Cincinnati Reefer View Post
    I had seen you recommend this on some other posts and decided to try it. It has now been soaking for about 12 hours. I will pull it out tomorrow and dry it and try again. I will update this post with my results. Thank you for replying zombie.
    And don't forget to float your calibration packet. That bites a lot of people in the butt.

    Sent from my SM-G965U using Tapatalk

Similar Threads

  1. Question: Quick Q - can we rename optical sensor output from "open/closed" to "wet/dry"?
    By BTBarney in forum Fluid Monitoring Module (FMM), FMK, ATK, LDK, and FMM Accessories
    Replies: 3
    Last Post: 07-07-2020, 16:49
  2. Help! Salinity Probe Calibration "Out of Range"
    By BillVH in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 1
    Last Post: 02-06-2019, 14:36
  3. Replies: 5
    Last Post: 07-15-2018, 10:12
  4. Solved: Virtual Outlet stuck "ON" when switched to "Auto"
    By Hondaman99 in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 2
    Last Post: 11-22-2015, 10:19
  5. Question: Temp probe "gone wild"? 101+ & email alarm.....
    By jimmerpro in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 6
    Last Post: 10-15-2014, 04:36

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •