    Temperature compensation on probes with new model Apex

    I have one of the new model Apex units. In the probe setup for pH and salinity, there are options for turning on the temperature compensation, which seem to have defaulted to off. In the old apex, you only wanted to turn them on if you were using something like a PM2 and had both the temp and pH probe in the same unit. But for just the base apex (new) unit, because you do have both in the same unit, do I assume right that you want them on?

  2. #2
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,464
    The use of TC is optional - your choice. I use it for both pH and conductivity (the latter at 2.2). The key thing is that if you have multiple temp probes, the “closest” temp probe must be in the same body of water at the pH or conductivity probe.

    By “closest”, I mean the temp probe connected to the same hardware unit as the pH or conductivity probe, whether it be an Apex base, PM1, or PM2.

    TC defaults to disabled rather than enabled because Neptune cannot assume that the temp probe will always be enabled and placed in the same body of water.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

Similar Threads

  1. Temperature Compensation for Salinity
    By demariners in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 8
    Last Post: 05-14-2020, 13:53
  2. PM2 temperature compensation?
    By jrizo1 in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 20
    Last Post: 10-03-2017, 09:13
  3. Question: Temperature Compensation
    By gpwdr in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 1
    Last Post: 12-23-2016, 21:38
  4. Question: Temperature Compensation
    By gpwdr in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 1
    Last Post: 12-23-2016, 21:38
  5. Can you specify the temperature probe that will be used for PH Temperature Compensation
    By Dxpert in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 2
    Last Post: 04-10-2015, 18:26

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •