    Programming Help, Unique scenario

    I am not sure if I have this in the right place. If not, I apologize in advance.

    I just purchased my first Apex system and am trying to learn the programming end. I do not use this system to control a fish tank. I am a manufacturer and I am using this system in our R&D facility to monitor pH and temp, turn on and off pumps, and to add certain chemicals to our treatment system using DOS. I have the system programmed based on what I have read in this forum however I am hoping to change the way it functions.

    Right now I have the DOS #1 set to the following:
    Set OFF
    If pH > 4.40 Then ON
    If pH < 4.30 Then OFF

    There are problems with the above though. Our system has 750 gallons of water. Our circulation pump turns all this volume every 15 minutes. By the time the pH changes to 4.30 and shuts the DOS off, there is too much acid added to the system. Over the next 5 minutes, the pH falls to 3.5. The DOS is typically adding around 200mL of acid at a time, which is too much. I have moved the DOS input as close to the probe as possible.

    What I would like, if possible:
    When pH hits 4.40 the DOS turn on and pumps 25ml of acid. Then the system waits for 15 minutes before cycling again. If after 15 minutes the pH is still above 4.40, the DOS pumps another 25ml of acid.

    There is no risk of 25ml of acid ever being too much to add. Is this possible? if so, can someone detail walk me through how to achieve this?

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Create a DOS profile that will dispense 25ml at the slowest rate, set the interval to 900 seconds and set the count to whatever you are comfortable with adding before you think it might become a problem (like maybe the pH probe failed or something) replace ON in your code with the profile name.

    Sent from my SM-G965U using Tapatalk

  3. #3
    New User
    Join Date
    Dec 2018
    Location
    Alberta
    Posts
    3
    Quote Originally Posted by zombie View Post
    Create a DOS profile that will dispense 25ml at the slowest rate, set the interval to 900 seconds and set the count to whatever you are comfortable with adding before you think it might become a problem (like maybe the pH probe failed or something) replace ON in your code with the profile name.

    Sent from my SM-G965U using Tapatalk
    Thank you. I am looking at the profile page and see PF 1 - 32. I assume I can change any one of them as they are not in use.

    If I use PF1
    Name: DOS_Control
    Type: DOS
    Dose Rate: 60mL/Min
    Direction: Forward
    Per DOse Amount: 25ml
    Dose intervals: 900
    Dose Count: 1 (I do not understand what you mean when you say what you are comfortable adding. Are you saying after it does 1 count it will no longer dose until ???)

    DOS Config
    Set OFF
    If pH > 4.40 Then DOS_Control

  4. #4
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    I would adjust to something like this

    If I use PF1
    Name: DOS_Control
    Type: DOS
    Dose Rate: 14mL/Min
    Direction: Forward
    Per DOse Amount: 25ml
    Dose intervals: 900
    Dose Count: 10

    DOS Config
    If pH > 4.40 Then DOS_Control
    If pH < 4.35 Then OFF


    This will run hysteresis and will very slowly add the acid over 2 minutes, which would likely help the stability. If pH is still above 4.35, the profile will try again 13 minutes later. The dose count acts as a maximum number of retrys before it stops trying. You could be alerted when this happens using the following format (example is for 10 retrys). You could then investigate the issue and reset it by setting the DOS to OFF and then back to AUTO.

    DOSfail
    Set OFF
    If Output DOSname = ON Then ON
    Defer 150:00 Then ON

    Email
    (Existing)
    If Output DOSfail = ON Then ON


    Sent from my SM-G965U using Tapatalk

  5. #5
    New User
    Join Date
    Dec 2018
    Location
    Alberta
    Posts
    3
    Quote Originally Posted by zombie View Post
    I would adjust to something like this

    If I use PF1
    Name: DOS_Control
    Type: DOS
    Dose Rate: 14mL/Min
    Direction: Forward
    Per DOse Amount: 25ml
    Dose intervals: 900
    Dose Count: 10

    DOS Config
    If pH > 4.40 Then DOS_Control
    If pH < 4.35 Then OFF


    This will run hysteresis and will very slowly add the acid over 2 minutes, which would likely help the stability. If pH is still above 4.35, the profile will try again 13 minutes later. The dose count acts as a maximum number of retrys before it stops trying. You could be alerted when this happens using the following format (example is for 10 retrys). You could then investigate the issue and reset it by setting the DOS to OFF and then back to AUTO.

    DOSfail
    Set OFF
    If Output DOSname = ON Then ON
    Defer 150:00 Then ON

    Email
    (Existing)
    If Output DOSfail = ON Then ON


    Sent from my SM-G965U using Tapatalk
    Thank you. This is working well. Last question. The DOSFAIL command you linked, does that go into the programming under "If pH < 4.35 Then OFF"?

  6. #6
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by RRand View Post
    Thank you. This is working well. Last question. The DOSFAIL command you linked, does that go into the programming under "If pH < 4.35 Then OFF"?
    No. DOSfail would be a virtual outlet and you throw that last line in your email alarm to get alerted.

    Sent from my SM-G965U using Tapatalk

Similar Threads

  1. Question: Power Fail Scenario - Need Clarification
    By TheNore in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 5
    Last Post: 10-24-2019, 09:47
  2. Question: Radion programming issue for Advanced programming vs. lighting wizard
    By Afm32607 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 1
    Last Post: 07-08-2017, 18:03
  3. Question: Radion programming issue for Advanced programming vs. lighting wizard
    By Afm32607 in forum Apex Programming for Lighting
    Replies: 1
    Last Post: 07-08-2017, 18:03

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •