    My 2 ph probes are way diff

    I did them both in the same calibration fluids separately of corse. Ones reading 7.95 one is 8.02 I just added a new one cause I’m setting up a calcium reactor any ideas or thoughts on this would be helpful!


    Sent from my iPhone using Tapatalk

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    If you didn't let them sit in the solution long enough each, that can cause skewing between the two. There is also non-linearity error in the pH probes that is in the 0.05 pH error vicinity.

    Sent from my SM-G965U using Tapatalk

  3. #3
    Frequent Visitor
    Join Date
    Jul 2015
    Location
    Los Angeles
    Posts
    51
    Quote Originally Posted by zombie View Post
    If you didn't let them sit in the solution long enough each, that can cause skewing between the two. There is also non-linearity error in the pH probes that is in the 0.05 pH error vicinity.

    Sent from my SM-G965U using Tapatalk
    Thanks so much for the reply. But when the calibration settled while calibrating I let them sit for 15-20 seconds. Is there a point in letting them sit longer? I guess my biggest concern is the .7 apart a big deal when using a calcium reactor?


    Sent from my iPhone using Tapatalk

  4. #4
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by LAreefguy View Post
    Thanks so much for the reply. But when the calibration settled while calibrating I let them sit for 15-20 seconds. Is there a point in letting them sit longer? I guess my biggest concern is the .7 apart a big deal when using a calcium reactor?


    Sent from my iPhone using Tapatalk
    They should sit for a minimum of 5 minutes. You can tell they are done similar to popcorn. If the reading has not changed in about 30 seconds it's done.

    Sent from my SM-G965U using Tapatalk

  5. #5
    Frequent Visitor
    Join Date
    Jul 2015
    Location
    Los Angeles
    Posts
    51
    Ok great soooooo last question. 5 min in the 7 and 5 min in the 10? I just checked the reading they are now .5 apart. Should I just be happy with that?


    Sent from my iPhone using Tapatalk

  6. #6
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by LAreefguy View Post
    Ok great soooooo last question. 5 min in the 7 and 5 min in the 10? I just checked the reading they are now .5 apart. Should I just be happy with that?


    Sent from my iPhone using Tapatalk
    Yes 5 min in each. I have had some probes take up to 10 minutes though. The number fully settling is the driver.

    Sent from my SM-G965U using Tapatalk

  7. #7
    Regular Vistor
    Join Date
    Jul 2018
    Location
    Adelaide, South Australia
    Posts
    24
    .I want to run a second Neptune ph probe from the Orp outlet. Is this possible & how, please?

  8. #8
    Master Control Freak RussM's Avatar
    Join Date
    Dec 2012
    Location
    California - US Pacific
    Posts
    22,464
    Quote Originally Posted by tuna01 View Post
    .I want to run a second Neptune ph probe from the Orp outlet. Is this possible & how, please?
    Using the Input setup page of the Classic Dashboard, enable pH2. This will also disable ORP.
    Please do not send me PMs with technical questions or requesting assistance - use the forums for Apex help. PM me ONLY if the matter is of a private or personal nature. Thanks.

Similar Threads

  1. Question: Configuring three diff pumps
    By downset21 in forum Apex Programming for Vortech, Tunze, & Other Controllable Pumps
    Replies: 17
    Last Post: 04-22-2018, 13:19
  2. VarSpd ports showing diff from apex2016 base than from VDM
    By lactose in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 1
    Last Post: 12-01-2016, 08:51
  3. Help! Module name diff in Fusion and classic
    By Wheeler89 in forum APEX Fusion
    Replies: 0
    Last Post: 03-21-2015, 13:53
  4. Diff between Apex, Lite and Jr
    By faulk28 in forum Pre-Sales Questions
    Replies: 1
    Last Post: 12-09-2013, 09:59
  5. Question: Does Apex support basic math? Like calc diff between temp probes?
    By jbraslins in forum Misc Apex Usage & Programming
    Replies: 1
    Last Post: 08-07-2013, 20:10

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •