I did them both in the same calibration fluids separately of corse. Ones reading 7.95 one is 8.02 I just added a new one cause I’m setting up a calcium reactor any ideas or thoughts on this would be helpful!
Sent from my iPhone using Tapatalk
I did them both in the same calibration fluids separately of corse. Ones reading 7.95 one is 8.02 I just added a new one cause I’m setting up a calcium reactor any ideas or thoughts on this would be helpful!
Sent from my iPhone using Tapatalk
If you didn't let them sit in the solution long enough each, that can cause skewing between the two. There is also non-linearity error in the pH probes that is in the 0.05 pH error vicinity.
Sent from my SM-G965U using Tapatalk
Thanks so much for the reply. But when the calibration settled while calibrating I let them sit for 15-20 seconds. Is there a point in letting them sit longer? I guess my biggest concern is the .7 apart a big deal when using a calcium reactor?
Sent from my iPhone using Tapatalk
Ok great soooooo last question. 5 min in the 7 and 5 min in the 10? I just checked the reading they are now .5 apart. Should I just be happy with that?
Sent from my iPhone using Tapatalk
.I want to run a second Neptune ph probe from the Orp outlet. Is this possible & how, please?