Thread: conductivity probe value shows different than what is shown

  1. #1
    Regular Vistor
    Join Date
    Apr 2013
    Location
    Portland, OR
    Posts
    20

    conductivity probe value shows different than what is shown

    Hi,
    I replaced my conductivity probe back in 7/11/18 due to the previous one stop measuring the salinity level. I had the original one for about 3 - 4 years.

    When I received a new probe, I perform the calibration for salt water and adjusted my programming to the setting as reported. It has always reported values of around 10.5 for 1.025 SG.

    When the value of the SG falls below 1.020, it reports values in the 90's. Recently it reported 95.5 for 1.019 SG. Now that I have restored the SG to 1.0.25, the app and Apex Fusion reports 10.8.

    When I set an alarm for Cond > 95.0, the e-mail alarm is triggered, which I don't think it should, since I'm seeing 10.8. So I'm suspecting the actual value is 105.8. I say this because when I set a rule for Cond < 11.0, I do not get an alarm, but the value is 10.8. So I should get an alarm.

    Here's my current log for 2/4:
    Item Date Status Statement
    1 2019-02-04 10:39:09 -0800 ON If Cond > 96.0 Then ON (10.8)
    2 2019-02-04 10:42:40 -0800 OFF Set OFF
    3 2019-02-04 10:43:21 -0800 ON If Cond > 99.9 Then ON (10.8)
    4 2019-02-04 10:45:16 -0800 OFF Set OFF

    Is the probe malfunctioning? Looking for suggestions to fix the issue.

    Jim

  2. #2
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Sounds like you didn't push the jack in all the way and then calibrated with the same settling value for both steps. Recalibrate the probe. It should read about 33 ppt for 1.025 SG.

    Sent from my SM-G965U using Tapatalk

  3. #3
    Regular Vistor
    Join Date
    Apr 2013
    Location
    Portland, OR
    Posts
    20
    I'll check the jack when I get home, but have tried to reseat it and did hear the clip click. My last probe read 33 ppt before.

    Thanks
    Jim

  4. #4
    Frequent Visitor
    Join Date
    Mar 2015
    Location
    San Jose,CA
    Posts
    528
    There is no clip for the cond probe. You should not hear any clicks.

  5. #5
    Regular Vistor
    Join Date
    Apr 2013
    Location
    Portland, OR
    Posts
    20
    you are correct.. I was thinking about the temp probe. Will check the connection and recalibrate.

Similar Threads

  1. Several apex users salinity probe suddenly shows 0 simultaneously
    By CaliReefer in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 6
    Last Post: 05-08-2022, 20:46
  2. EB4 Amp 'Probe' shows old Aquabus module Probe name
    By Mal7887 in forum EnergyBars 832 / 632 / 8 / 6 / 4
    Replies: 1
    Last Post: 06-10-2018, 10:02
  3. Non existing PH Probe shows readings
    By flyairbus in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 3
    Last Post: 05-31-2018, 07:40
  4. Graph history shows for only ONE Ph probe...
    By GREEF stricken in forum Apple iOS App for APEX Fusion
    Replies: 4
    Last Post: 11-24-2015, 19:17
  5. Conductivity Probe - Conducity Range set to Salinity (PPT or Conductivity) ?
    By omcnair in forum Apex Classic/Apex Lite/Apex Jr
    Replies: 2
    Last Post: 12-25-2014, 19:31

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •