Check your water line connections on the manifold behind the reagent tray!
So after connecting and going thru the start up instructions I could not get my Trident to function due to error of - NO WATER SAMPE- I rechecked all lines and no kinks or obstructions. Then I moved sample line to a different area where it didn't have to pass thru any stand or sump partitions, and yet NO WATER SAMPLE. Finally I slide the reagent tray out completely and found that the incoming water line looked to be connected but was actually not seated fully onto the nipple of the manifold. There was however a nice puddle of liquid coming from behind the Trident at this point after the machine trying to pull sample water from the tank several times during diagnostics, since the tube was going thru the dose head but not connected to the manifold the sample water and some blueish liquid had puddled and drained out. Glad that I hadn't fully installed the Trident onto the shelf in the sump where this may have drained into the water since Im guessing the blueish liquid was one of the reagents!
Just my personal encounter, and thought this may help someone searching the forum.