Page 2 of 2 FirstFirst 12
Results 26 to 38 of 38

Thread: pH probe reading low ???

  1. #26
    New User
    Join Date
    Apr 2013
    Location
    Phoenix
    Posts
    7
    Another update. Just cross-checked against a new hand-held pH meter (Dr Meter, pH 100) calibrated at 6.86, and it's basically spot on. Apex reads 7.77, Dr Meter 7.8. Apex reads the calibration buffer as 6.86. I guess I have to take it all back ;-)

  2. #27
    Regular Vistor
    Join Date
    Dec 2013
    Location
    Tucson,AZ
    Posts
    46
    I don't have a meter to check, but I did order another probe. It's only 7$ and from China so we will see how that works out,


    Sent from my iPhone using Tapatalk

  3. #28
    New User
    Join Date
    Aug 2016
    Location
    Alpharetta, GA
    Posts
    3
    Just an update. Purchased another pH probe (BRS one) and hooked it to the Apex. pH readings now match across all my devices.

  4. #29
    Regular Vistor
    Join Date
    Dec 2013
    Location
    Tucson,AZ
    Posts
    46
    Got probe for 7$. China. Don't do it.


    Sent from my iPhone using Tapatalk

  5. #30
    New User
    Join Date
    May 2017
    Location
    Dubai
    Posts
    2
    After reading the entire thread, i am sure something is not right with Apex, since i also get the same reading of about 7.72 at night and 7.78-7.80 in the afternoon, further upon checking with API and sali test kits, my ph always come 8.20 to 8.40. I have though calibrated the probe just once, dont know what to do, i am thinking to calibrate the probe once again. I think that the placement of probe also matters, since i have placed all my probes directly into the tank near the return just to get accurate reading from the tank itself. Any suggestions on how to go about it

  6. #31
    New User
    Join Date
    May 2016
    Location
    Fort Myers, FL
    Posts
    4
    I have had similar experiences. I am running two lab grade probes on apex. (Brs and neptune). Multiple calibrations with pinpoint solutions. And they run about .3 below seneye. Tried moving to different locations in tank with no affect in difference. This is over multiple seneye slides. I have not checked into calibrating seneye or adjusting it.

    Sent from my SM-T530NU using Tapatalk

  7. #32
    New User
    Join Date
    May 2016
    Location
    Fort Myers, FL
    Posts
    4
    I am using apex classic (gold)

    Sent from my SM-T530NU using Tapatalk

  8. #33
    New User
    Join Date
    Jun 2017
    Location
    Fairfax, VA
    Posts
    3
    My probe has been reading the same a well. .3 exactly my Apex (brand new) shows 7.8 and my Seneye (brand new) shows 8.1. Has anyone been told that there is an issue with the probe? I have been running my system for almost 2 months now and would like some clarity.

  9. #34
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Calibrate your probe again and make sure to let the value settle (I wait 5 minutes) and make sure there are no bubbles in the probe during the process. After calibration is complete verify that the probe reads 7.0 and 10.0 in each solution. If it doesn't recalibrate with the same solution. If it does, the apex value will be within 0.02 of the actual lab measured pH at any point between 7 and 10.

    If that reading after proper calibration doesn't jive with another test kit, Seneye, etc. Check that the apex probe reading is the same in the tank compared to a cup of tank water isolated from the tank. If that reading is identical, you can be certain the apex is right and the secondary measurement is wrong. If the reading changes, you have an electrical interference or stray voltage issue that must be found and eliminated.

    You might be an engineer if...You have no life and can prove it mathematically.

  10. #35
    NSI Member Krazie4Acans's Avatar
    Join Date
    Nov 2013
    Location
    US, Mountain Time
    Posts
    1,255
    The temp of the calibration solutions should also match the tank temp and be maintained during the calibration. Then once calibrated you can turn on temperature compensation which will help reduce reading swings due to temperature changes in the tank.

  11. #36
    New User
    Join Date
    Jun 2017
    Location
    Fairfax, VA
    Posts
    3
    Thank you Zombie- Thank you Krazie4Acans. I received an email from Apex and they instructed me to do the similar steps you guys suggested. I am a new user and used the auto calibration. I was instructed to use the manual calibration which I think is way more precise. I really let the probes sit for a bit. All is right with the world my PH bounces between 8.12 and 8.29 so I am very happy. now will re calibrate in 6 months.

  12. #37
    NSI Member Krazie4Acans's Avatar
    Join Date
    Nov 2013
    Location
    US, Mountain Time
    Posts
    1,255
    That's great to hear.

  13. #38
    Frequent Contributor zombie's Avatar
    Join Date
    Dec 2013
    Location
    Denver, CO
    Posts
    13,176
    Quote Originally Posted by Krazie4Acans View Post
    The temp of the calibration solutions should also match the tank temp and be maintained during the calibration. Then once calibrated you can turn on temperature compensation which will help reduce reading swings due to temperature changes in the tank.
    While this is terribly important for conductivity, the effect is nearly neglegable for pH because the calibration solutions are temperature buffered and pH probes are not highly temperature dependent. In even extreme cases the error is likely to be less than 0.05. With that said, if you are determined to have absolute pin-point accuracy then the solution should match tank temp. Personally I don't bother with this step as I value my time more than 0.05 pH error.

    You might be an engineer if...You have no life and can prove it mathematically.

Page 2 of 2 FirstFirst 12

Similar Threads

  1. Salinity probe reading low
    By Nlutfi in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 0
    Last Post: 03-31-2022, 12:50
  2. Question: Salinity Probe Reading LOW
    By inqztve in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 28
    Last Post: 02-20-2018, 09:27
  3. pH Reading Low
    By mikes781 in forum AquaBus Modules, Probes, and Breakout Boxes
    Replies: 5
    Last Post: 12-07-2017, 11:24
  4. Question: pH reading low
    By Js.Aqua.Project in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 2
    Last Post: 11-21-2016, 12:59
  5. salinity probe reading low after water change
    By hamel.aquarium in forum A2 Apex/ApexEL and A3 Apex Pro/Apex/Apex Jr
    Replies: 3
    Last Post: 09-30-2016, 07:21

Bookmarks

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •